Measuring Low-Ohm Resistors: The Four-Terminal Kelvin Method

A concern we regularly get from customers is low-value resistors that seem to be out of specification when measured. Ranging from simple ten or one-ohm parts down to mili-ohm or lower shunt resistors, customers often test these products the same way the test more standard four-digit resistance devices and discover them measuring well outside their listed specs. Is this normal, and are the parts really bad?

Often, no. Standard two-wire measurements are often insufficient to ascertain the value of low-resistance products, as lead resistance, contact resistance, and other imprecisions in the measurement set-up or equipment can overwhelm the low resistance values of these products. Instead, set-ups for measuring these products use what is often called the four-wire Kelvin method. This sensing method supplies current through a device from two terminals while sensing voltage over the part through a second set of terminals, eliminating lead and contact resistance. A circuit layout for such sensing can be seen here:


4-terminal sensing circuit layout, courtesy of Wikipedia and Erik Buer

This layout will better allow you to measure and verify low-resistance devices.

For more information on the four-terminal method, and on resistors in general, check our or Resistor engineering brief here: Resistors.

For more information, please refer to Measuring Low Resistance values below 0.1 OHM