The main difference is probe length, per the datasheet:
OXY-FLEX-0: 220 mm probe
OXY-FLEX-1: 400 mm probe
Everything else is the same: the 24 V supply, the outputs selectable via a jumper on the pcb (4-20 mA, 0-10 V or RS232), the 0.1-25% or 0.1-100% O₂ range, and the standard -100°C to +250°C gas temperature limit. Adding “-H” to either one (for example OXY-FLEX-0-H) raises the gas temperature limit to +400°C.
The longer -1 probe is useful when the probe must reach farther through a thick wall, insulation or a large duct. The shorter -0 suits thinner walls or tighter spaces. The datasheet’s mounting drawing also shows the housing staying in atmospheric air at 60°C max while the probe tip sits in the hot gas, so the right length depends on how your installation separates the two.