Vacuum suitability and outgassing test results of 5TC-TT-T-40-72 thermocouples

------Question for 5TC-TT-T-40-72 Please Put your question below------

Hi,

We (Isaware Oy) are interested in using these thermocouples in thermal balance and thermal vacuum tests of space instrument for ESA. However, as this is space flight hardware, there are strict cleanliness requirements on any equipment going inside the vacuum chamber with the instrument. According to the ECSS standard, the equipment shall be tested for outgassing in standard conditions:

  • Duration: 24 hours
  • Temperature: 125 °C on the material sample
  • Collector Plate Temperature: 25 °C (to catch condensing volatiles)
  • Vacuum Pressure: ≤ 10⁻³ Pa

and the acceptance limits are:

  • RML (Recovered Mass Loss): < 1.0% (includes water reabsorption)
  • CVCM (Collected Volatile Condensable Material): < 0.10%

Have these thermocouples, including the cables, been tested for vacuum use? Is there an outgassing test report available? Some kind of documentation proving that the thermocouples are compliant with these requirements is required before they can be approved by ESA.

Hi @arto.lehtolainen,

I have sent your request to Omega. I will update you when they reply.

Best regards.

Hi @arto.lehtolainen,

Omega told us they have no outgassing test data for this part, including the cable. Since the cable insulation is PFA, they suggested asking DuPont, the PFA maker, if they have data for their material.