I am working on a Water Flow Test Stand at the University of Southern California. We are considering the ifm SU8621 flow sensor for installation in a nominal 1-inch water line and would appreciate your guidance on calibration and suitability for our application.

Our system uses compressed air to pressurize a water tank, driving water through a control valve, past a pressure measurement point, and through a test article. Only water flows through the sensor. We use volumetric flow rate and the pressure drop across the test article to calculate its flow coefficient, Cv.

Currently, we calculate flow from the change in collected water mass measured by load cells. We would like to introduce real-time flow measurement and improve accuracy. Because we have concerns about the accuracy of our existing load cell measurements, we are looking for a way to calibrate and validate the SU8621 against an independent reference.

What calibration method would you recommend for this application? Would timed water collection using a separate calibrated scale be suitable, or would you recommend a calibrated reference flowmeter? We would also appreciate guidance on how to apply any calibration correction to the SU8621 and what accuracy we could reasonably expect after calibration.

Could you also clarify the installation requirements, including straight pipe lengths and placement relative to the control valve, and whether you offer calibration services or recommend a calibration provider?

We would be happy to provide a plumbing diagram and additional operating details if helpful. Thank you for your time and guidance.

Greetings,

The operation instructions for SU8621 contain installation recommendations on or about page 11.

Measurements of fluid flow rate are known for being difficult to establish to a high precision; it’s the sort of business where figures having more than 2 significant digits warrant skepticism.

A bucket and stopwatch have long been the gold standard for validation of such things where feasible. Rates are ephemeral, volumes less so.

One may expect post-calibration accuracy to be no better than the repeatability of the measurement, mentioned as 0.2% in the datasheet. As for how one might develop such a calibration, the typical process is to take measurements spanning a variable of interest (flow rate would seem an obvious first choice) compare readings to a reference standard, and fit a curve mapping one to the other. If the process involves more than one variable, the problem then becomes multi-dimensional.

If the load cells used to measure mass presently are of questionable accuracy, it might be worth investigating those as an alternative/parallel approach. “Real time flow rate” is a somewhat malleable concept, and given a sufficiently fast signal chain one could imagine a 1-second average being a close-enough approximation for most purposes.

In addition to coming up with your own validation tests you might want to buy the ISO calibration certificate option for the sensor, it seems reasonably priced.